Graeme Gill wrote:

> I did a simple test using my ColorMunki and a CRT based monitor.
> I've had the instrument taped to the one spot, and the display
> and instrument powered up for some hours. There was a noticeable
> amount of drift over the first hour or two, but after that the
> readings were fairly constant.

Thanks for testing!
I would consider the temperature of a CRT tube as very cool (isolating
vacuum), compared to a LCD screen with backlight, so I'm really
astonished that even with a CRT there is such a long lasting drift.

> Using spotread, the error to a reading taken as a reference once the drifting
> had finished remains under 1 delta E, and averages about 0.5 delta E.
> The smallest error is less than 0.1 delta E.

The smaller temperature difference (room temperature<->CRT) should
result in a lower drift, so this does not signify that your
ColorMunki is better/less prone to drift, right?

> Recalibrating does not seem to affect this.

This means that your ColorMunki is also not back to its first readings
after a self-calibration?

> [I'd guess that the variation in this case is due to the lack of
> synchronization between the CRT refresh rate and the instrument integration 
> time.]

This reminds me of the question, if it is possible that the temporal
dithering used in panels to gain 10 Bit resolution is jointly
responsible for measurement variance?

Regards, Juergen