Hi Johnny,

In most VNAs you have to create a calibration Standard for each TRL, and most 
VNAs support as many lines as needed for their Bandwidth, i.e. if your VNA only 
supports 3GHz it is unlikely to support >4 Line Standard.
You should look for an application note of the instrument to find out how to 
define a cal. Standard. In some instruments this is very unhandy and you have 
to write .mdf files including all TRL data at the correct position. Other VNA's 
come with a TRL whizzard.

BR
Gert


----------------------------------------
Absender ist HARTING Electronics GmbH, Marienwerderstraße 3, D-32339 Espelkamp; 
Registergericht: Amtsgericht Bad Oeynhausen; Register-Nr.: HRB 8808; 
Vertretungsberechtige Geschäftsführer: Dipl.-Kfm. Edgar-Peter Düning, 
Dipl.-Ing. Torsten Ratzmann, Dipl.-Wirtschaftsing. Ralf Martin Klein

-----Ursprüngliche Nachricht-----
Von: FU, Yejun [mailto:yfu@xxxxxxxxx]
Gesendet: Donnerstag, 5. Dezember 2013 15:45
An: Havermann, Gert
Cc: si-list@xxxxxxxxxxxxx
Betreff: RE: [SI-LIST] AW: TRL Calibration Question

Hello Gert:

        No doubt about the reflection, I will check the LOAD impedance.  For 
the second one, the problem is this VNA only have three line option for 
lines(1,2,3) but actually the calibration provide four info of Delay line 1, 
line 2, line3 , and Load. I believe that the load is used to provide DC to a 
low frequency response.  My third question, they provide a calibration line at 
this board, I wasn't sure how to check it, I will try it by checking the phase 
continuities.  For the SMAs, I know its won't support 30GHZ. But their put the 
this range on their board, my results also shows that it's done a good job I 
mean as good as 2.92mm, but I don't the reason.

Best Regard,

Johnny

-----Original Message-----
From: si-list-bounce@xxxxxxxxxxxxx [mailto:si-list-bounce@xxxxxxxxxxxxx] On 
Behalf Of Havermann, Gert
Sent: Thursday, December 05, 2013 3:35 AM
To: si-list@xxxxxxxxxxxxx
Subject: [SI-LIST] AW: TRL Calibration Question

Hi Johnny,

Answer 1) What will create more reflections: LOAD or OPEN? You should use the 
one with mor reflections if the reflect is measured Answer 2) You need 
different lengths of thru to cover a greater frequency range. The measurement 
per line is only valid within a 180° phase change (due to the fact that phase 
is directly related to length, but the phase keeps repeating itself while the 
length increases endlessly. That's why you have to be inside the 180° range to 
have meaningful results) Answer 3) You should check the Source match using an 
Airline. Also a sanity check with measuring a straigt line and check if the 
phase has any discontinuities. Attach a Load and check the reflection noise 
floor level.

B.T.W.: no regular SMA will give you 30GHz Bandwidth. Most SMAs end at 18GHz, 
there are some available up to 26GHz. For everything above that there are other 
coax connectors available (e.g. K, 2.92mm)

BR
Gert


----------------------------------------
Absender ist HARTING Electronics GmbH, Marienwerderstraße 3, D-32339 Espelkamp; 
Registergericht: Amtsgericht Bad Oeynhausen; Register-Nr.: HRB 8808; 
Vertretungsberechtige Geschäftsführer: Dipl.-Kfm. Edgar-Peter Düning, 
Dipl.-Ing. Torsten Ratzmann, Dipl.-Wirtschaftsing. Ralf Martin Klein

-----Ursprüngliche Nachricht-----
Von: si-list-bounce@xxxxxxxxxxxxx [mailto:si-list-bounce@xxxxxxxxxxxxx] Im 
Auftrag von Yejun Fu
Gesendet: Donnerstag, 5. Dezember 2013 02:14
An: si-list@xxxxxxxxxxxxx
Betreff: [SI-LIST] TRL Calibration Question

Hi All:
          I have this calibration board, has an OPEN(80mils), THUR(160mils), 
LOAD, 3 different lines(1,2,3),  1x line, 2x line, calibration verification 
line .  The calibration board also gives info of delay for trace of load, 
line1, line2, line3, I believe it covers from DC to 30GHz, (SMA connectors). 
This calibration board is used to evaluate a high speed connector(DUT), which 
has a two paddle card with a 80 mils transmission line on header and receptacle 
side.  So my DUT structure is SMA+80mils
trace+high speed connector+80 mils trace+SMA.

My understand is this is a TRL calibration board to put my reference plane to 
the pins of high speed connector.  My VNA has an option called multi-section 
TRL, which ask me to do the Reflection, Thur, line (3 different line) only.

I have three questions:

1. For VNA's reflection option, which trace on the calibration board should I 
use, I mean OPEN or LOAD?

2. Why we need line 1 line2 line3  three different lines for TRL?

3. How to check my calibration accuracy?

Thanks

Johnny


------------------------------------------------------------------
To unsubscribe from si-list:
si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field

or to administer your membership from a web page, go to:
//www.freelists.org/webpage/si-list

For help:
si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field


List forum  is accessible at:
               http://tech.groups.yahoo.com/group/si-list

List archives are viewable at:
                //www.freelists.org/archives/si-list

Old (prior to June 6, 2001) list archives are viewable at:
                http://www.qsl.net/wb6tpu


------------------------------------------------------------------
To unsubscribe from si-list:
si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field

or to administer your membership from a web page, go to:
//www.freelists.org/webpage/si-list

For help:
si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field


List forum  is accessible at:
               http://tech.groups.yahoo.com/group/si-list

List archives are viewable at:
                //www.freelists.org/archives/si-list

Old (prior to June 6, 2001) list archives are viewable at:
                http://www.qsl.net/wb6tpu


------------------------------------------------------------------
To unsubscribe from si-list:
si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field

or to administer your membership from a web page, go to:
//www.freelists.org/webpage/si-list

For help:
si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field


List forum  is accessible at:
               http://tech.groups.yahoo.com/group/si-list

List archives are viewable at:     
                //www.freelists.org/archives/si-list
 
Old (prior to June 6, 2001) list archives are viewable at:
                http://www.qsl.net/wb6tpu
  

Other related posts:

  • » [SI-LIST] AW: AW: TRL Calibration Question - Havermann, Gert