[optimal] Re: Topcon error codes and misfiring

  • From: angiolith@xxxxxxx
  • To: optimal@xxxxxxxxxxxxx
  • Date: Tue, 20 Sep 2011 15:21:08 -0400 (EDT)

Subject: Re: optimal Digest V2 #189



If the problem is only FA related, it is most likely related to the filter 
solenoids.   Sounds like one of the solenoids is sticking or failing.  But here 
is where you need to do further differential diagnosis.  

1. Does the problem occur if you leave BOTH filters in for the entire 
angiogram? You can run a dummy angiogram or 2 to test this.  If the camera 
works, the solenoids or switches are obviously the problem.
2.You might try leaving the exciter filter in and letting the camera insert the 
barrier filter automatically for a few FA's.  If you get no errors, reverse the 
process. This should narrow down which filter solenoid is failing.  It should 
also offer you a temporary fix  (i.e., shoot FA's in the mode that works until 
you can get a repair done).  If you get failures in both modes, the problem may 
not be the solenoids themselves but the circuit board that triggers them, or a 
sensor that waits until the filters are seated to fire the camera.

The other camera is a bit trickier.  Your description implies that the trigger 
is not sending a signal to start the photographic sequence (send relay mirror 
up, fire flash, grab picture to computer), but I would check all the 
connections (fundus camera to relay lens, digital camera to computer, anything 
else you can jiggle).  The wires in the camera that run the joystick trigger 
are small, and the welds can break, but if that is the case, it won't be long 
before an intermittent problem becomes a permanent one.

If all else fails, call Topcon.

Marty Rothenberg
Angiographics, Inc.

 

 


 

 

-----Original Message-----
From: FreeLists Mailing List Manager <ecartis@xxxxxxxxxxxxx>
To: optimal digest users <ecartis@xxxxxxxxxxxxx>
Sent: Tue, Sep 20, 2011 1:09 am
Subject: optimal Digest V2 #189


optimal Digest  Mon, 19 Sep 2011        Volume: 02  Issue: 189

In This Issue:
                [optimal] Topcon error codes and misfiring
                [optimal] Re: Topcon error codes and misfiring
                [optimal] Re: Topcon error codes and misfiring

----------------------------------------------------------------------

From: "Breit, Peter" <BreitP@xxxxxxxx>
Date: Mon, 19 Sep 2011 11:19:19 -0400
Subject: [optimal] Topcon error codes and misfiring

Hello colleagues,
                        I am having problems with the IVFA filters on the 
Topcon 
TRC-50DX .
The camera suddenly stops working and I am getting all kind of error codes 
(3,6,8,10..). The first 2-3 IVFAs are fine but then the problem acts up. Maybe 
it is heat related ?
I tried different settings, manual filter insertion, etc. but it does not help. 
Letting the camera rest, turned off for � hour seems to help.. but obviously I 
don't  have the time for that. Obviously it is only IVFA related.
Maybe someone had the same problem and can tell me how it was solved. New 
solenoid put in ?  My local Topcon rep could not fix it.

Incidentally another Topcon in a different office is also misfiring. Meaning 
sometimes you have to press the joystick button several times before you can 
hear the mirror move up and a picture appears. I may get several blank (black) 
pictures on the screen and then a good fundus picture..  I am not sure if the 
problem is in the joystick or the eye piece head ?
Again this doesn't happen right away, usually on the 3rd or 4th IVFA.... This 
happens not only when doing IVFA but also when just shooting fundus 
photographs...

Any help is very much appreciated.

Thank you
Peter

Peter L. Breit, CRA.
Director Ophthalmic Services
The Lankenau Hospital & Bryn Mawr Hospital
Philadelphia, PA 19096

484-476-3338
484-476-8206 fax
484-437-3262 mobile
page 3707








------------------------------

Date: Mon, 19 Sep 2011 12:59:57 -0400
From: "Egnatz,  Thomas James" <tegnatz@xxxxxxxxx>
Subject: [optimal] Re: Topcon error codes and misfiring

Peter,

Not sure if this might be related, but check all the relay lens and 
digital digital back to computer connections.  On ours, I had a problem 
a couple of months age:  pressing the joystick button caused the flash 
to fire, and it sounded like a regular good exposure, but a black image 
would appear on the monitor.  Turns out the cable from the relay lens 
was plugged into a computer cable, but the connecters were not screwed 
into each other.  The slightest tension on one would cause the above 
problem.

I don't recall if there was an error code.

One thing to add to your troubleshooting list.

Tom





Quoting "Breit, Peter" <BreitP@xxxxxxxx>:

> Hello colleagues,
>                        I am having problems with the IVFA filters on
> the Topcon TRC-50DX .
>
> The camera suddenly stops working and I am getting all kind of error
> codes (3,6,8,10..). The first 2-3 IVFAs are fine but then the problem
> acts up. Maybe it is heat related ?
> I tried different settings, manual filter insertion, etc. but it does
> not help. Letting the camera rest, turned off for ? hour seems to
> help.. but obviously I don't  have the time for that. Obviously it is
> only IVFA related.
> Maybe someone had the same problem and can tell me how it was solved.
> New solenoid put in ?  My local Topcon rep could not fix it.
>
> Incidentally another Topcon in a different office is also misfiring.
> Meaning sometimes you have to press the joystick button several times
> before you can hear the mirror move up and a picture appears. I may
> get several blank (black) pictures on the screen and then a good
> fundus picture..  I am not sure if the problem is in the joystick or
> the eye piece head ?
> Again this doesn't happen right away, usually on the 3rd or 4th
> IVFA.... This happens not only when doing IVFA but also when just
> shooting fundus photographs...
>
> Any help is very much appreciated.
>
> Thank you
> Peter
>
> Peter L. Breit, CRA.
> Director Ophthalmic Services
> The Lankenau Hospital & Bryn Mawr Hospital
> Philadelphia, PA 19096
>
> 484-476-3338
> 484-476-8206 fax
> 484-437-3262 mobile
> page 3707
>
>
>
>
>
>
>




------------------------------

From: David <daveclark86@xxxxxxxxx>
Subject: [optimal] Re: Topcon error codes and misfiring
Date: Mon, 19 Sep 2011 15:47:11 -0400

Peter,
  Did your troubleshooting fix your camera?  If not let me know and I will ask 
around for you. 

Warm Regards,

Dave Clark, COA

Regional Sales Manager
US Dealers Partners 
Topcon Medical Systems
111 Bauer Drive, Oakland, NJ 07436
Dclark@xxxxxxxxxx
Cell: 551-579-1062

On Sep 19, 2011, at 12:59 PM, "Egnatz,  Thomas James" <tegnatz@xxxxxxxxx> wrote:

> Peter,
> 
> Not sure if this might be related, but check all the relay lens and digital 
digital back to computer connections.  On ours, I had a problem a couple of 
months age:  pressing the joystick button caused the flash to fire, and it 
sounded like a regular good exposure, but a black image would appear on the 
monitor.  Turns out the cable from the relay lens was plugged into a computer 
cable, but the connecters were not screwed into each other.  The slightest 
tension on one would cause the above problem.
> 
> I don't recall if there was an error code.
> 
> One thing to add to your troubleshooting list.
> 
> Tom
> 
> 
> 
> 
> 
> Quoting "Breit, Peter" <BreitP@xxxxxxxx>:
> 
>> Hello colleagues,
>>                       I am having problems with the IVFA filters on
>> the Topcon TRC-50DX .
>> 
>> The camera suddenly stops working and I am getting all kind of error
>> codes (3,6,8,10..). The first 2-3 IVFAs are fine but then the problem
>> acts up. Maybe it is heat related ?
>> I tried different settings, manual filter insertion, etc. but it does
>> not help. Letting the camera rest, turned off for ? hour seems to
>> help.. but obviously I don't  have the time for that. Obviously it is
>> only IVFA related.
>> Maybe someone had the same problem and can tell me how it was solved.
>> New solenoid put in ?  My local Topcon rep could not fix it.
>> 
>> Incidentally another Topcon in a different office is also misfiring.
>> Meaning sometimes you have to press the joystick button several times
>> before you can hear the mirror move up and a picture appears. I may
>> get several blank (black) pictures on the screen and then a good
>> fundus picture..  I am not sure if the problem is in the joystick or
>> the eye piece head ?
>> Again this doesn't happen right away, usually on the 3rd or 4th
>> IVFA.... This happens not only when doing IVFA but also when just
>> shooting fundus photographs...
>> 
>> Any help is very much appreciated.
>> 
>> Thank you
>> Peter
>> 
>> Peter L. Breit, CRA.
>> Director Ophthalmic Services
>> The Lankenau Hospital & Bryn Mawr Hospital
>> Philadelphia, PA 19096
>> 
>> 484-476-3338
>> 484-476-8206 fax
>> 484-437-3262 mobile
>> page 3707
>> 
>> 
>> 
>> 
>> 
>> 
>> 
> 
> 
> 
> 

------------------------------

End of optimal Digest V2 #189
*****************************


 
 

Other related posts: