Jim Ferwerda wrote:

Hi,

1) In v 2.1.1 spotread -a works ok, but dispread doesn’t recognize the -a flag

yes, since dispread -a support was added in V2.1.2.

2) I updgraded to v 2.1.2 and now both spotread -a and dispread -a work, but 
dispread was intermittently failing with the error

at first the failures seemed random so I suspected a cabling or usb 
communication problem, but the problem persisted even after cable swaps and 
reconnections. I then noticed that the failures appeared to be correlated 
with the “brightest” patches in the .ti1 file (e.g. full R,G,B channels and 
W).

The CM has a relatively limited maximum brightness limits due to its minimum
integration time. (I suspect that there is no good technical reason for this, 
but
X-Rite perhaps wanted to make some space between the CM and the i1Pro).

3) To test the hypothesis that things were failing due to the big numbers, I 
put a 0.6 (-2 stop) ND filter in front of the ambient sensor, and everything 
ran fine. I also did tests using a 0.3 (-1 stop) ND filter (failures) and a 
combination of 0.3 and 0.15 filters (-1.5 stops, only failures on full green 
and full white), so this seems to confirm the hypothesis.

Right.

- Is this a known issue with measuring bright things using the CM and/or 
dispread?

AFAIK it's a limitation of the CM.

- Do you have recommendations for how to best characterize my system? 
(ColoMunki, DLP projector set up for small/bright FOV)

I think you've found the simplest/cheapest solution - ND filters, assuming they
really are neutral.
Another solution would be be to switch to an i1Pro, or a spectrometer
with a much wider measurement range such as the JETI 1211.

Cheers,
        Graeme Gill.