Martin Weberg wrote:
2010/2/8 Graeme Gill <graeme@xxxxxxxxxxxxx>:
Martin Weberg wrote:
For completeness, here we go again using v1.1.0. Notice the bump in
plot 2 (St. dev. at 410nm is 0.2 (n=3)).
It would be interesting to know how the bump is created, but
without that paper to measure it is difficult to track down.

Something I could provide, besides the paper?

Probably the spectrum of the paper measured without the UV filter
will be enough.

readings contain no UV (10-400nm) I thought it was interesting that
the Spectrolino fitted with UV filter produced higher reflectance then
without filter below 420nm.

On thinking about this some more, I can see only two possible
explanations:

1) There is something different about the instrument calibration
   for UV vs. no UV filter. Measuring the calibration tile may
   reveal whether this is the case.

2) There's some level of up-conversion fluorescence, or similar
   going on. If it really is up-conversion fluorescence then this
   would be difficult to model, since it is typically non-linear
   (two-photon absorption),
   The "or similar" would have to be something like the high level
   of illumination at the fluorescent emission wavelengths suppressing
   absorption at shorter wavelengths. Fluorescence saturation
   may behave like this, but it is also non-linear and typically
   only happens at very high illumination levels.

Graeme Gill.