[argyllcms] Re: Is the i1D3 (ColorMunki Display) accurate?

  • From: Claas Bickeböller <lists@xxxxxxxxxxxxxxxxx>
  • To: argyllcms@xxxxxxxxxxxxx
  • Date: Fri, 27 Jul 2012 13:47:09 +0200

Dear Stephen,

maybe this report gives you an idea of the interinstrument and intermodel 
disagreement you can expect from typical low cost sensors.

http://www.fogra.org/dokumente/upload/6297b_dis_aymeasurement_intermediate_reportv3.pdf

Best regards

Claas


Am 27.07.2012 um 13:28 schrieb Stephen T:

> The i1D2 is less than two years old. I've kept it well stored, with silica 
> gel desiccant.
>  
> Stephen.
>  
> From: "János, Tóth F." <janos666@xxxxxxxxxx>
> To: argyllcms@xxxxxxxxxxxxx 
> Sent: Friday, 27 July 2012 8:27 PM
> Subject: [argyllcms] Re: Is the i1D3 (ColorMunki Display) accurate?
> 
> How old is your i1d2?
> 
> 2012/7/27 Stephen T <stwebvanuatu@xxxxxxxxxxxx>
> Hello Janos,
>  
> Thankyou for your constructive advice.
>  
> There are two issues: 1) accuracy and 2) unit-to-unit variability:
>  
> http://www.drycreekphoto.com/Learn/Calibration/MonitorCalibrationHardware.html
>  
> 
> OEM i1D2 variability DE00 = 4 - 7, average white point accuracy DE00 = 2 - 4
> i1D3 variability DE00 = 0.4 - 2, average white point accuracy DE00 = 2 - 3
>  
> In the worst case, any two instruments might disagree by DE00 = 2 + 7 = 9.
> My results show a  difference of DE00 = 12.
> I also measured white patches with spotread and calculate DE94 = 14.
> I suppose that luminance measurements are fairly robust and my i1D3 does 
> appear to be an odd unit in two tests.
>  
> I think my i1D3 is measuring out of tolerance?
> Has anyone else experienced similar and have they measured the difference?
>  
> I'm not throwing my Spyder2 in the bin just yet.
>  
> Stephen.
>  
> From: "János, Tóth F." <janos666@xxxxxxxxxx>
> To: argyllcms@xxxxxxxxxxxxx 
> Sent: Friday, 27 July 2012 7:00 AM
> Subject: [argyllcms] Re: Is the i1D3 (ColorMunki Display) accurate?
> 
> I think you bought a new sensor because you believed it's better than your 
> old. Didn't you?
> 
> The i1d3 is more accurate in theory. The hardware is better, the corrections 
> are smarter.
> Of yourse, it's not "perfect". (Neither any other relatively cheap 
> instruments are.) But I guess you will be happy with the i1d3 if the i1d2 was 
> good enough for your needs.
> 
> 2012/7/26 Stephen T <stwebvanuatu@xxxxxxxxxxxx>
> Hello,
> 
> I received my ColorMunki display this week and have been doing some 
> comparisons. It gives different readings to my OEM i1D2.
> 
> I calibrated my wide gamut NEC P221W display with X-Rite's ColorMunki Display 
> software and the following settings:
> native white point (hardware setting), D65, gamma = 2.2, brightness 120 cd/m2 
> (adjusted in hardware), WGCCFL display type (under preferences!), no ambient 
> adjustment, no flare correction.
> 
> I then measured the response in Argyll CMS v1.4.0 with the profile loaded in 
> Windows 7. I measured with two instruments: ColorMunki Display and NEC 
> MDSVSENSOR (OEM i1D2). I also measured without the CCSS matrix. The profile 
> was good when measured with the i1D3 and so I believe Argyll CMS is not at 
> fault. The profile was inaccurate when tested with the i1D2.
> 
> Are the i1D3 generic correction matrices accurate? My MDSVSENSOR suggests the 
> white point DE00 = 12 and brightness is 8 cd/m2 brighter.
> 
> Has anyone else tested the accuracy of the i1D3 against other instruments, 
> especially spectrophotometer?
> 
> Are the generic correction matrices accurate for some display technologies 
> and inaccurate for others? I assume the MDSVSENSOR is accurate for wide gamut 
> NEC displays. I bought the ColorMunki Display mainly for standard gamut CCFL 
> and WLED displays and now I am hesitant to use it!
> 
> Maybe one of my colorimeters is faulty?
> 
> Stephen.
> 
> Here are the results:
> 
> i1D3 + CCSS 
> ----------- ---
>  
> C:\Users\Stephen>dispcal -d 1 -y l -X WGCCFLFamily_07Feb11.ccss -r -P 
> 0.5,0.5,2.0 
> Place instrument on test window. 
> Hit Esc or Q to give up, any other key to continue: 
> Current calibration response: 
> Black level = 0.12 cd/m^2 
> White level = 119.57 cd/m^2 
> Aprox. gamma = 2.18 
> Contrast ratio = 1004:1 
> White chromaticity coordinates 0.3108, 0.3274 
> White    Correlated Color Temperature = 6622K, DE 2K to locus =  4.8 
> White Correlated Daylight Temperature = 6622K, DE 2K to locus =  0.2 
> White        Visual Color Temperature = 6442K, DE 2K to locus =  4.6 
> White     Visual Daylight Temperature = 6614K, DE 2K to locus =  0.2 
> The instrument can be removed from the screen. 
>  
> OEM i1D2 (NEC MDSVSENSOR) 
> ------------------------- ------------
>  
> C:\Users\Stephen>dispcal -d 1 -y l -r -P 0.5,0.5,2.0 
> Place instrument on test window. 
> Hit Esc or Q to give up, any other key to continue: 
> Current calibration response: 
> Black level = 0.12 cd/m^2 
> White level = 127.53 cd/m^2 
> Aprox. gamma = 2.17 
> Contrast ratio = 1026:1 
> White chromaticity coordinates 0.3135, 0.3503 
> White    Correlated Color Temperature = 6328K, DE 2K to locus = 14.4 
> White Correlated Daylight Temperature = 6321K, DE 2K to locus = 12.0 
> White        Visual Color Temperature = 5810K, DE 2K to locus = 14.1 
> White     Visual Daylight Temperature = 5924K, DE 2K to locus = 11.6 
> The instrument can be removed from the screen.
> 
> i1D3 
> -----
>  
> C:\Users\Stephen>dispcal -d 1 -y l -r -P 0.5,0.5,2.0 
> Place instrument on test window. 
> Hit Esc or Q to give up, any other key to continue: 
> Current calibration response: 
> Black level = 0.12 cd/m^2 
> White level = 118.54 cd/m^2 
> Aprox. gamma = 2.17 
> Contrast ratio = 1002:1 
> White chromaticity coordinates 0.3013, 0.3325 
> White    Correlated Color Temperature = 7100K, DE 2K to locus = 12.4 
> White Correlated Daylight Temperature = 7090K, DE 2K to locus =  9.6 
> White        Visual Color Temperature = 6517K, DE 2K to locus = 12.0 
> White     Visual Daylight Temperature = 6666K, DE 2K to locus =  9.3 
> The instrument can be removed from the screen.
> 
> 
> 
> 
> 
> 
> 
> 


Other related posts: