[SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- From: "Cavanna, Vicente Vaca (Sr. ; ProCurve ASICs)" <vicente.cavanna@xxxxxx>
- To: <homer.si.mpson66@xxxxxxxxx>, <si-list@xxxxxxxxxxxxx>
- Date: Wed, 25 Apr 2007 00:20:02 -0000
With a VNA one can increase the power level of the stimulus and get
more energy into and through the DUT even when the DUT is lossy and/or
extremely reflective. With a TDR this is not as easy to do; in fact the
TDRs I have used do not permit such control.
Vicente
> -----Original Message-----
> From: si-list-bounce@xxxxxxxxxxxxx=20
> [mailto:si-list-bounce@xxxxxxxxxxxxx] On Behalf Of Peter J
> Sent: Tuesday, April 24, 2007 9:36 AM
> To: si-list@xxxxxxxxxxxxx
> Subject: [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
>=20
> Hi,
> The biggest problem in using a TDR or VNA for that matter is=20
> to get enough energy into the DUT since even a very small=20
> inductance at the connection point from the TDR to the DUT=20
> will give an insertion loss. It's possible to get a return=20
> loss of -20dB at 5GHz if you work hard. At 50 GHz it's hard=20
> to get -3 dB RL which means that half of the power will enter=20
> the DUT and the rest is wasted. By using a VNA it is possible=20
> to make a TRL calibration kit and remove the effect of the=20
> testsystem and only get the DUT characteristics.
> I use a short pigtail with a thin semirigid coax instead of a=20
> probe when I do measurements with the TDR(tektronix). It=20
> increases the accuracy a great deal.
>=20
> BR
> Peter
>=20
> -----Original Message-----
>=20
> From: si-list-bounce@xxxxxxxxxxxxx
> [*mailto:si-list-bounce@xxxxxxxxxxxxx*<si-list-bounce@xxxxxxxxxxxxx>]
> On Behalf Of ext Asbenson, Lyndell L
>=20
> Sent: 24. april 2007 02:27
>=20
> To: heidi.barnes@xxxxxxxxxx; si-list@xxxxxxxxxxxxx
>=20
> Cc: timothy.j.nash@xxxxxxxx; tom_cip_11551@xxxxxxxxx
>=20
> Subject: [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
>=20
> Dedicated TDR is much easer to setup then going the VNA route -Lyndell
>=20
> -----Original Message-----
>=20
> From: si-list-bounce@xxxxxxxxxxxxx
> [*mailto:si-list-bounce@xxxxxxxxxxxxx*<si-list-bounce@xxxxxxxxxxxxx>
> ]
>=20
> On Behalf Of Barnes, Heidi
>=20
> Sent: Monday, April 23, 2007 12:17 PM
>=20
> To: si-list@xxxxxxxxxxxxx
>=20
> Cc: timothy.j.nash@xxxxxxxx; tom_cip_11551@xxxxxxxxx
>=20
> Subject: [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
>=20
> Hi Tom,
>=20
> Tim is correct in that probing is not a simple task. If you=20
> have the time and patience, a Vector Network Analyzer with=20
> PCB Thru-Reflect-Line Calibration standards can provide a=20
> method for getting measured electrical data for a PCB=20
> structure that does not include the probes or connectorized=20
> transitions on to the printed circuit board. The S-Parameter=20
> data can then be converted to time domain for TDR analysis of=20
> the discontinuities and a 40GHz box will give you around an=20
> 18pS rise time pulse (an edge that is ~100mils long for a PCB=20
> dielectric constant of ~4) which gives a significant amount=20
> of detail (50mils is rather ambitious unless you have money=20
> for the top end 110GHz VNA's or 8pS rise time TDR scopes).=20
> The other thing to think about is that most applications=20
> running about 3Gbps are differential so it is also nice to=20
> have a 4-Port system for measurements.
>=20
> Regards,
>=20
> Heidi
>=20
> -----Original Message-----
>=20
> From: si-list-bounce@xxxxxxxxxxxxx
> [*mailto:si-list-bounce@xxxxxxxxxxxxx*<si-list-bounce@xxxxxxxxxxxxx>
> ]
>=20
> On Behalf Of Nash, Timothy J
>=20
> Sent: Monday, April 23, 2007 11:33 AM
>=20
> To: tom_cip_11551@xxxxxxxxx; si-list@xxxxxxxxxxxxx
>=20
> Subject: [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
>=20
> Hi, Tom.
>=20
> I have tried using a Tektronix TDR (I'm not sure of the=20
> model) to probe PCBs before, and found that what probe you=20
> were using, and WHAT you were probing, can make or break your=20
> measurement. We were using a hand-held pressure-actuated type=20
> probe and simply bending your wrist a little would change the=20
> response. We had a tough time even discerning between two=20
> vias spaced an inch apart from each other. I believe the=20
> probes used by most PCB manufacturers are on a robotic arm=20
> and they are probing PCB coupons that have pre-defined TDR=20
> features. I found it very difficult to glean much useful info=20
> from real PCB traces when the discontinuities were as close=20
> together as what you are describing. I don't believe this was=20
> a fault of the scope - just a function of the probing method=20
> and the feature constraints of the PCB.
>=20
> Tim
>=20
>=20
>=20
>=20
>=20
> -----Original Message-----
>=20
> From: si-list-bounce@xxxxxxxxxxxxx
> [*mailto:si-list-bounce@xxxxxxxxxxxxx*<si-list-bounce@xxxxxxxxxxxxx>
> ]
>=20
> On Behalf Of tom_cip_11551
>=20
> Sent: Monday, April 23, 2007 10:47 AM
>=20
> To: si-list@xxxxxxxxxxxxx
>=20
> Subject: [SI-LIST] TDR Analsis Equipment..Tektronix vs Agilent?
>=20
> Hi to the Signal Integrity community.
>=20
> I am looking for a set of equipment with wich to do TDR=20
> analysis on=3D3D20 high speed PCB structures. We are now=20
> building boards to run at=3D3D20 speeds of 3 Gb/s or faster and=20
> I would like to resolve=3D3D20 discontinuites within, say, 50=20
> mils to 80 mils of each other.
>=20
> I have gotten quotes from Tektronix and Agilent for TDR=20
> equipment. =3D3D20
>=20
> The Tektronix system consists of the DSA8200 digital sampling=20
> =3D scope,=3D3D20 the 80E04 TDR sampling module and the 80SICON=20
> signal integrity and=3D3D20 failure analysis software (a $20K=20
> item by itself).
>=20
> The Agilent system consists of the 86100C scope, the 54754A=20
> TDR=3D3D20 module and various other software programs for S=20
> parameters, Jitter=3D3D20 analysis, etc.
>=20
> I am wondering what the Signal Integrity community at large=20
> is=3D3D20 using.
> Which system is more popular and, most important, cost=3D3D20 =
effective?
>=20
> What are some of the advantages and drawbacks of each?
>=20
> Are there other solutions on the market that are more cost effective?
>=20
> Thank You
>=20
> Tom
>=20
>=20
>=20
> ------------------------------------------------------------------
>=20
> To unsubscribe from si-list:
>=20
> si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
>=20
> or to administer your membership from a web page, go to:
>=20
> *http://www.freelists.org/webpage/si-list*<http://www.freelist
s.org/webpage/si-list>
>=20
> For help:
>=20
> si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
>=20
>=20
>=20
> List technical documents are available at:
>=20
> *http://www.si-list.net* <http://www.si-list.net/>
>=20
> List archives are viewable at: =3D3D20
>=20
> *http://www.freelists.org/archives/si-list*<http://www.freelis
ts.org/archives/si-list>
>=20
> or at our remote archives:
>=20
> *http://groups.yahoo.com/group/si-list/messages*<http://groups
.yahoo.com/group/si-list/messages>
>=20
> Old (prior to June 6, 2001) list archives are viewable at:
>=20
> *http://www.qsl.net/wb6tpu* <http://www.qsl.net/wb6tpu>
>=20
> =3D3D20
>=20
> ------------------------------------------------------------------
>=20
> To unsubscribe from si-list:
>=20
> si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
>=20
> or to administer your membership from a web page, go to:
>=20
> *http://www.freelists.org/webpage/si-list*<http://www.freelist
s.org/webpage/si-list>
>=20
> For help:
>=20
> si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
>=20
>=20
>=20
> List technical documents are available at:
>=20
> *http://www.si-list.net* <http://www.si-list.net/>
>=20
> List archives are viewable at: =3D3D20
>=20
> *http://www.freelists.org/archives/si-list*<http://www.freelis
ts.org/archives/si-list>
>=20
> or at our remote archives:
>=20
> *http://groups.yahoo.com/group/si-list/messages*<http://groups
.yahoo.com/group/si-list/messages>
>=20
> Old (prior to June 6, 2001) list archives are viewable at:
>=20
> *http://www.qsl.net/wb6tpu* <http://www.qsl.net/wb6tpu>
>=20
> =3D3D20
>=20
> ------------------------------------------------------------------
>=20
> To unsubscribe from si-list:
>=20
> si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
>=20
> or to administer your membership from a web page, go to:
>=20
> *http://www.freelists.org/webpage/si-list*<http://www.freelist
s.org/webpage/si-list>
>=20
> For help:
>=20
> si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
>=20
>=20
>=20
> List technical documents are available at:
>=20
> *http://www.si-list.net* <http://www.si-list.net/>
>=20
> List archives are viewable at: =3D20
>=20
> *http://www.freelists.org/archives/si-list*<http://www.freelis
ts.org/archives/si-list>
>=20
> or at our remote archives:
>=20
> *http://groups.yahoo.com/group/si-list/messages*<http://groups
.yahoo.com/group/si-list/messages>
>=20
> Old (prior to June 6, 2001) list archives are viewable at:
>=20
> *http://www.qsl.net/wb6tpu* <http://www.qsl.net/wb6tpu>
>=20
> =3D20
>=20
> ------------------------------------------------------------------
>=20
> To unsubscribe from si-list:
>=20
> si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
>=20
> or to administer your membership from a web page, go to:
>=20
> *http://www.freelists.org/webpage/si-list*<http://www.freelist
s.org/webpage/si-list>
>=20
> For help:
>=20
> si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
>=20
>=20
>=20
> List technical documents are available at:
>=20
> *http://www.si-list.net* <http://www.si-list.net/>
>=20
> List archives are viewable at:
>=20
> *http://www.freelists.org/archives/si-list*<http://www.freelis
ts.org/archives/si-list>
>=20
> or at our remote archives:
>=20
> *http://groups.yahoo.com/group/si-list/messages*<http://groups
.yahoo.com/group/si-list/messages>
>=20
> Old (prior to June 6, 2001) list archives are viewable at:
>=20
> *http://www.qsl.net/wb6tpu* <http://www.qsl.net/wb6tpu>
>=20
>=20
> ------------------------------------------------------------------
> To unsubscribe from si-list:
> si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
>=20
> or to administer your membership from a web page, go to:
> http://www.freelists.org/webpage/si-list
>=20
> For help:
> si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
>=20
>=20
> List technical documents are available at:
> http://www.si-list.net
>=20
> List archives are viewable at: =20
> http://www.freelists.org/archives/si-list
> or at our remote archives:
> http://groups.yahoo.com/group/si-list/messages
> Old (prior to June 6, 2001) list archives are viewable at:
> http://www.qsl.net/wb6tpu
> =20
>=20
>=20
------------------------------------------------------------------
To unsubscribe from si-list:
si-list-request@xxxxxxxxxxxxx with 'unsubscribe' in the Subject field
or to administer your membership from a web page, go to:
http://www.freelists.org/webpage/si-list
For help:
si-list-request@xxxxxxxxxxxxx with 'help' in the Subject field
List technical documents are available at:
http://www.si-list.net
List archives are viewable at:
http://www.freelists.org/archives/si-list
or at our remote archives:
http://groups.yahoo.com/group/si-list/messages
Old (prior to June 6, 2001) list archives are viewable at:
http://www.qsl.net/wb6tpu
- Follow-Ups:
- [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- From: Tom Dagostino
- References:
Other related posts:
- » [SI-LIST] TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- » [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- [SI-LIST] Re: TDR Analsis Equipment..Tektronix vs Agilent?
- From: Tom Dagostino